Plotting concentration in ppm on the x-axis and intensity in mV on the y-axis producesa linear...
Plotting concentration in ppm on the x-axis and intensity in mV on the y-axis producesa linear calibration of y-125x +254. (a) What is the sensitivity? (2 pts) (b) What is the detection limit if the standard deviation of the lowest concentration is 1114 mV? (3 pts) (c) What is the concentration of an unknown that produced a signal of 6998 mV? (3 pts) Calculate the S/N ratio for the following peaks: peak 1 intensity 5400, peak 2 intensity 650, and peak 3 intensity 25400. The noise is 110.(use whole numbers, no decimals) (6 pts)