6a) A tumor located at the depth of 10cm is treated using a
non-isocentric technique with a Co60 machine, to a dose of 1.8 Gy,
with a single 5cm x 5cm beam. Compute the maximum dose received
anywhere in the patient's anatomy. Refer to the attached table for
PDD data.
b) At what depth does the maximum dose occur? Explain.
c) Compute the dose at a depth of 5ccm along the beam axis.
d) Would this be a good treatment strategy? Explain.
Attached Table:
Field Size (cm2) |
5 x 5 |
10 x 10 |
20 x 20 |
|
Depth (cm) |
||||
0.5 |
100 |
100 |
100 |
|
5.0 |
76.7 |
80.4 |
83.0 |
|
10.0 |
53.3 |
58.7 |
63.3 |
Answer:
For a),b)& c)
Relative dose measurements:
It is generally performed in a large water phantom and for this
purpose a suitable ionization
chamber is used in the water phantom. In this way dose values are
determined at many
points under varying treatment conditions, such as field size,
source skin distance (SSD),
and the presence of beam modifiers[3].
When multiple fields are used for treatment of a particular tumor
inside the patient,
isocentric, source axis distance (SAD) setups are often used. In
contrast to source skin
distance SSD setups, which rely on percentage depth dose (PDD)
distributions, SAD setups
rely on other functions, such as tissue-air ratios and
tissue-phantom ratios, for dosimetric
calculations. So acquiring a group of measurements is divided to
two groups the first group
is the central axis depth doses in water (SSD SET-UP) and the
second group is the central
axis depth doses in water (SAD SET-UP).
Concerning the central axis depth doses in water (SSD SET-Up) the
measuring sessions of
the PDD is to place the center of the ion chamber in the water
phantom over the range of
depth of 0 to 25 cm in increments not larger than 1 cm and over the
range of field sizes
rf .
from 4cm x 4 cm to 40cm x 40cm in increments not larger than 5 cm
for the side. The
central axis dose distributions inside the phantom are usually
normalized to depth of
maximum D max = 100%.
The measurements of the central axis depth doses in water for (SAD
SET-UP) consist of
the tissue-air ratio (TAR), tissue-phantom ratio (TPR), tissue
maximum ratio (TMR),
output factor, blocking tray factor, and beam profiles.
Concerning the measuring sessions of the (TAR) (suggested by Johns
in 1953), the ion
chamber should be placed in water phantom over the range of same
depths and field size as
PDD measurements, in both water and air. The ratio between the
absorbed dose in water at
the same point in free air is referred to as the (TAR). [7]
The measurements of the TPR at a point in a water phantom
irradiated by a photon beam
can be done by divide the total absorbed dose at that point by the
total absorbed dose at a
point or the beam axis at affixed reference depth, usually 5 cm or
10 cm.
i
TMR is special case of TPR and may be determined by dividing the
dose at a given point in
water phantom to the dose at the same point at the reference depth
of maximum dose[7].
Absolute dose measurements:
Absolute dose measurements are also performed in a water
phantom. For this purpose a
calibrated ionization chamber is used. The generally applied
procedure for specification of
beam quality using IAEA TRS-398 protocol is to specify the
tissue-phantom ratio,
TPR20,10. This is the ratio of the absorbed doses at depths of 20
cm and 10 cm in a water
phantom, measured with a constant source-chamber distance of 100 cm
and a field size of
10x10 cm2 at the plane of the chamber [2].
-• J& Determination of absorbed dose to water under the
reference conditions is given' by
pf'-
multiplying the corrected reading of the dosimeter with the
calibration factor arid KQ go
which is the chamber-specific factor which corrects for the
difference between the
reference beam quality go and the actual quality being used Q [
Answer d) When the tumor size is within the above mentioned limit this is a good treatment to be followed.
6a) A tumor located at the depth of 10cm is treated using a non-isocentric technique with...
A) a tumour located at a depth of 10cm is treated using a non-isocentic technique with a Co-60 machine, to a dose of 1.8Gy, with a single 5cm x 5cm beam. compute the maximum dose received anywhere in the patient's anatomyrefer to table for PDD data depth cm field size --> 5 x 5 10 x 10 20 x 20 0.5 100 100 100 5.0 76.7 80.4 83.0 10.0 53.3 58.7 63.3 B) at what depth does the maximum dose...
.6MeV beam, to a tumour located at a depen of 13 cm. An 11 × 11 cm neld is to be used with . 100 cm SSD. For this beam in this etry, you 12x12cm field at depth of 10cm . 689%, POD for a 10x10cm feld at depth n, even the sewne aantona, data POO fur ,10.10cm, ne4 at depth of 10 cm . 676% 15 cm-52.6%, and POD for a 12 x 12 om feid at depth of...