a.
F = IqI v B
where,
q = 1.6 x 10^-19 (this is the charge of proton )
v = 0.15 m/s
B = 0.22 T
F = 1.6 x 10^-19 x 0.22 x 0.15
F = 5.28 x 10^21 N
b.
F = qE
E = F/q
E = (5.28 x 10^-21)/ (1.6*10^-19)
E = 0.033 V/m = 3.3 x 10^-2 N/C
c.
3.3 × 10−2 V/m x 0.0034 m = 1.112 x 10^-4 V
Typical blood velocities in the coronary arteries range from 10 to 30 cm/s. An electromagnetic flowmeter...
Typical blood velocities in the coronary arteries range from 10 to 30 cm/s. An electromagnetic flowmeter applies a magnetic field of 0.21 T to a coronary artery with a blood velocity of 12 cm/s. As we can see in the figure below, this field exerts a force on ions in the blood, which will separate. The ions will separate until they make an electric field that exactly balances the magnetic force. This electric field produces a voltage that can be...
An electromagnetic flowmeter is to be used to measure blood speed. A magnetic field of 0.145 T is applied across an artery of inner diameter 3.80 mm. The Hall voltage is measured to be 89.8 V. What is the average speed of the blood flowing in the artery? cm/s
You are working on an electromagnetic method for measuring the speed at which blood flows through arteries and veins inside the human body. Blood contains both positive and negative ions flowing together in the same direction. If those currents are exposed to a perpendicular magnetic field, they will be pushed in opposite directions, causing a charge separation between the two sides of the blood vessel. But the separated charges create an electric field that inhibits charge separation, so the effect...
Blood contains positive and negative ions, which experience a magnetic force in a magnetic field as they move. An Electromagnetic Flowmeter takes advantage of this phenomenon. A heart surgeon monitors the flow rate of blood through an artery. Electrodes make contact with the outer surface of the blood vessel to measure an induced voltage. For a magnetic field strength of 0.037 T, a potential difference of 160 µV appears between the electrodes for a blood vessel of inside diameter 2.9...