the following samples were analyzed for silver. to each digestion flask 200 ml aliquots of ten different drinking water samples were added. silver was extracted by a co-precipitation procedure and collected by filtration. the precipitate was digested in 200 microliters hot concentrated nitric acid and diluted to 5 ml in a dilute phosphate solution. standards were prepared according to the following table. the samples were analyzed by graphite furnace AAS and compared to the calibration curve. results of the analyses follow: (chart in picture)
*** calculate the concentration of silver in each sample. are there any problems? how will you address each of the identified problems?
ans)
from above data that
The model:
A = 0.0077*C
or
C = A/0.0077
Then... substitute this equation for each sample:
NO | A | C |
2 | 0.079 | 10.3 |
3 | 0.058 | 7.5 |
4 | 0.129 | 16.8 |
5 | 0.082 | 10.6 |
6 | 0.797 | 103.5 |
7 | 0.063 | 8.2 |
8 | 0.009 | 1.2 |
9 | 0.028 | 3.6 |
10 | 0.136 | 17.7 |
The only problem is that at high concentrations, such as ppm > 100 the model starts to fail, but this should not be of much attention since concentrations are not above 100 ppb