A lithium ion-selective electrode gave the potentials given in the table below for the following standard...

A lithium ion-selective electrode gave the potentials given in the table below for the following standard solutions of LiCl and two samples of unknown concentration

Solution ([Li+]/M)

Potential vs. SCE, mV

0.100 M

+1.0

0.050 M

-30.0

0.010 M

-60.0

0.001 M

-138.0

Unknown 1

-48.5

Unknown 2

-75.3

  1. Draw a calibration curve of the electrode potential versus log[Li+] and determine if the electrode is displaying nernstain behaviour.
  2. Use a linear least squares procedure to determine the concentration of the unknowns
0 0
Add a comment Improve this question Transcribed image text
Answer #1

Solution ([Li+]/M)

Potential vs. SCE, mV

log[Li+]

[Li+]/M

for

Unknown

0.100

+1.0

-1

-

0.050

-30.0

-1.30

-

0.010

-60.0

-2

-

0.001

-138.0

-3

-

Unknown 1

-48.5

-1.7

(Calculated from plot)

0.02

Unknown 2

-75.3

-2.1

(Calculated from plot)

0.008

0 -20 -40 (-1.7, -48.5) -60 -80 (-2.1, -753) -100 -120 -140 -2.2 -1.4 -2.6 -2.4 -2.0 -1.8 -1.6 -1.0 -3.0 -2.8 -1.2 log[Li] Po According to Nernst equation,

RT In(Li nF E E

So, E - constant x log(Li E X

If we plot electrode potential(E) versus log[Li+], the data points are fitted with straight line. So, given electrode is displaying nernstain behaviour.

From the plot the concentration of Unknown 1 and Unknown 2 are 0.02 (M) and 0.008 (M) respectively.

Add a comment
Know the answer?
Add Answer to:
A lithium ion-selective electrode gave the potentials given in the table below for the following standard...
Your Answer:

Post as a guest

Your Name:

What's your source?

Earn Coins

Coins can be redeemed for fabulous gifts.

Not the answer you're looking for? Ask your own homework help question. Our experts will answer your question WITHIN MINUTES for Free.
Similar Homework Help Questions
  • Calibration data for a bromide ion‑selective electrode (ISE) was collected and recorded in the table. The...

    Calibration data for a bromide ion‑selective electrode (ISE) was collected and recorded in the table. The potential of the ISE was measured against a saturated calomel electrode (SCE). All solutions were buffered at a pH of 7.56. A linear calibration curve can be constructed from this data as a plot of potential (in mV) vs. pBr. Determine the slope and y‑intercept of such a plot of the calibration data given. Br-M Potential (mV 0.300 -21.6 0.0300 36.7 0.00300 96.5 0.000300...

  • Calibration data for a bromide ion-selective electrode (ISE) was collected and recorded below. The potential of...

    Calibration data for a bromide ion-selective electrode (ISE) was collected and recorded below. The potential of the ISE was measured against a saturated calomel electrode (SCE). All solutions were buffered at a pH of 7.56. A linear calibration curve can be constructed from this data as a plot of potential (in mV) vs. pBr. Determine the slope and y-intercept of such a plot of the calibration data given [Br] (M) Potential (mV) 0.300 0.0300 0.00300 96.4 0.000300 155.1 0.0000300 224.8...

  • (5) The following data represent the potentials of a fluoride ion-selective electrode vs SCE for several...

    (5) The following data represent the potentials of a fluoride ion-selective electrode vs SCE for several solutions. All solutions are made up in a way to control their pH and ionic strength, and contain added EDTA. Prepare an excel plot of the measured potential vs pF based on this information and determine the concentration of fluoride in each unknown solution. 5 Solution (M) Potential (mv) 5.0 x 102 -22.4 .0 x 103 36.8 5.0 x 104 96.0 5.0 x 10-...

  • Potential (mV) [Br](M) 0.300 Calibration data for a bromide ion-selective electrode (ISE) was collected and recorded in...

    Potential (mV) [Br](M) 0.300 Calibration data for a bromide ion-selective electrode (ISE) was collected and recorded in the table. The potential of the ISE was measured against a saturated calomel electrode (SCE). All solutions were buffered at a pH of 7.56. A linear calibration curve can be constructed from this data as a plot of potential (in mV) vs. pBr. Determine the slope and y-intercept of such a plot of the calibration data given. -21.2 0.0300 36.2 95.7 0.00300 0.000300...

  • Question 4 of 4 > Potential (my -21.9 Calibration data for a bromide ion-selective electrode (ISE) was collected...

    Question 4 of 4 > Potential (my -21.9 Calibration data for a bromide ion-selective electrode (ISE) was collected and recorded in the table. The potential of the ISE was measured against a saturated calomel electrode (SCE). All solutions were buffered at a pH of 7.56. A linear calibration curve can be constructed from this data as a plot of potential (in mV) vs. pBr. Determine the slope and y-intercept of such a plot of the calibration data given. [Br"](M) 0.300...

  • A sodium ion electrode was calibrated by measuring the cell potentials of several standard Na+ solutions....

    A sodium ion electrode was calibrated by measuring the cell potentials of several standard Na+ solutions. The other electrode was the SCE. The calibration data are shown in the following table. Log[Na+] Ecell (mV) -1 -165.6 0 -110.0 1 -51.8 2 6.5 3 64.5 a. Plot a calibration curve for these data. b. Determine the slope and intercept for the regression line of this plot. Is the response of this sodium electrode Nernstian? Explain. c. A 10.00mL aliquot of table...

ADVERTISEMENT
Free Homework Help App
Download From Google Play
Scan Your Homework
to Get Instant Free Answers
Need Online Homework Help?
Ask a Question
Get Answers For Free
Most questions answered within 3 hours.
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT