A satisfactory radiographic image was obtained using 300 mA, 20ms at 90 kVp, 182cm SID, 20cm OID, 1mm FSS, 8:1 grid, 400 IP (CR), and no pathological condition.
In order to maintain the Detective Quantum Efficiency (DQE) of the image, what would be the new exposure factor at 91 cm?
A satisfactory radiographic image was obtained using 300 mA, 20ms at 90 kVp, 182cm SID, 20cm...
Radiology 9:26 Х Student Portal 1. A satisfactory radiograph is obtained using 10 mAs at an SID of 40". If the technologist were to repeat the exposure at a new SID of 72", what mAs value would be required to maintain radiographic exposure? 2. If a radiographic image is obtained at 40" using 32 mAs, and a repeat exposure would be performed at an ID of 80", what mAs would be required to produce an image demonstrating similar exposure? 3....