Mean value =(sum of the values)/(no of values taken)
Here, mean =(0.639+0.638+0.640+0.639+0.640+0.639+0.638)/7
= 4.473/7
= 0.639
According to the question, the expected value is 0.640
The difference between two absorption values = 0.640-0.639
= 0.001
Percentage error = difference between two values * 100/ actual value
= 0.001*100/0.640
= 0.156%
So, the difference between expected value and measured value is very less. Hence it is acceptable.
[ In practical situation, this deviation is not a big deal for UV-visible instrument. Sometimes instrumental error comes in this range or fluctuations is observed. ]
To test a spectrophotometer’s accuracy a solution of 60.06 ppm K2Cr2O7 in 5.0 mM H2SO4 is...
1. What is the %T for a sample if its absorbance is 1.27? 2. The accuracy of a spectrophotometer is evaluated by preparing a solution of 60.06 ppm K2Cr2O7 in 0.0050 M H2SO4, and measuring its absorbance at a wavelength of 350 nm in a cell with a pathlength of 1.00 cm. The expected absorbance is 0.640. What is the expected molar absorptivity of K2Cr2O7 at this wavelength?
help please? this was the only other information given REPORT SHEET Determination of the Solubility-Product Constant for a Sparingly Soluble Salt EXPERIMENT 8 A. Preparation of a Calibration Curve Initial (Cro121 0.0024 M Absorbance 5 mL Volume of 0.0024 M K Cro Total volume 1. I mL 100 mL 2. 100ML 3. 10 mL 100ml 4. 15 mL 100 ML Molar extinction coefficient for [CrO2) [Cro,2) 2.4x100M 12x1044 2.4810M 3.6810M 0.04) 2037.37 0.85 1.13 2. 3. Average molar extinction coefficient...