X-RAY CT
a) Consider a 20 cm path through water, with a small piece of
bone of thickness x. Assuming water has an attenuation coefficient
of 0.3 cm-1, and bone of 1.0 cm-1, how thick would the bone have to
be to reduce the number of photons detected by 10%?
b) Partial volume effects are troublesome when imaging some
portions of the brain. Illustrate where partial volume effects lead
to loss of tissue visualization. Suggest three independent ways to
potentially reduce this effect by altering the way imaging is
performed.
The total length is 20+x cm so, distance traveled in water is: 20 cm
The attenuation follows the exponential decay given by:
here, I = [1 - 0.1] Io
=>
taking natural log on both sides gives,
=> x = - 6.105 cm.