The calibration of a pressure sensor with a range of 0-16 bar is checked by applying known inputs to it in steps of 2 bar. The following measurements were obtained.
Known input pressure (bar) |
0 |
2 |
4 |
6 |
8 |
10 |
12 |
14 |
16 |
Output reading (V) |
0.00 |
0.55 |
1.25 |
1.70 |
2.35 |
3.15 |
3.65 |
4.05 |
4.85 |
MATLAB CODE:
x = 0:2:16; % known input pressures
y = [0.00 0.55 1.25 1.70 2.35 3.15 3.65 4.05 4.85]; % output
values
plot(x,y,'bo','MarkerSize',6);
[p,S] = polyfit(x,y,1); % best fit line
[y_fit,delta] = polyval(p,x,S); % error estimates
hold on
plot(x,y_fit,'r-');
grid;
legend('Actual data','Line of best fit');
clc;
fprintf('The sensitivity is %f ',p(1));
fprintf('The maximum error is %f percent ',delta(1)*100/4.85);
OUTPUT:
The sensitivity is 0.301250 The maximum error is 2.365051 percent
The calibration of a pressure sensor with a range of 0-16 bar is checked by applying known inputs...
2.21 The calibration of a spring balance is checked by applying a set of known masses to it, with the following results: Known input (g) 10 20 30 40 50 60 70 80 Output reading (g) 15.5 25.0 32.0 45.0 52.0 61.0 74.5 81.0 Draw a graph by plotting the data points on the graph paper provided and draw a best-fit straight line through the data points Which of the following static measurement characteristics are evident in the input-output graph...
summatize the following info and break them into differeng key points. write them in yojr own words apartus 6.1 Introduction—The design of a successful hot box appa- ratus is influenced by many factors. Before beginning the design of an apparatus meeting this standard, the designer shall review the discussion on the limitations and accuracy, Section 13, discussions of the energy flows in a hot box, Annex A2, the metering box wall loss flow, Annex A3, and flanking loss, Annex...