A set of samples of known concentration were used to create a calibration curve for rubxanthin. The data are shown below. Calibration Curve for Rubixanthin 1.60 1.40 a 1.20 1.00 0.80 0.60 0.40 0.20 0...
A set of samples of known concentration were used to create a calibration curve for rubxanthin. The data are shown below. Calibration Curve for Rubixanthin 1.60 1.40 a 1.20 1.00 0.80 0.60 0.40 0.20 0.00 y 114800x R2 0.9999 0.00E+00 2.00E-06 4.0OE-06 6.00E-06 8.00E-06 1.00E-05 1.20E-05 1.40E-05 Concentration (M) 30) An extract of rubixanthin of unknown concentration was analyzed with UV-Vis Spectroscopy. The unknown had an absorbance of 0.83. What is the concentration (M) of rubixanthin in the unknown?
A set of samples of known concentration were used to create a calibration curve for rubxanthin. The data are shown below. Calibration Curve for Rubixanthin 1.60 1.40 a 1.20 1.00 0.80 0.60 0.40 0.20 0.00 y 114800x R2 0.9999 0.00E+00 2.00E-06 4.0OE-06 6.00E-06 8.00E-06 1.00E-05 1.20E-05 1.40E-05 Concentration (M) 30) An extract of rubixanthin of unknown concentration was analyzed with UV-Vis Spectroscopy. The unknown had an absorbance of 0.83. What is the concentration (M) of rubixanthin in the unknown?