Choose two of the following tests/equipment below, tell us what test(s) can be performed and provide written instructions on how to operate and maintain the equipment.
Be sure to include the following information and instructions in your submission:
A) Cleaning the exterior:
1) First, disconnect the Hemocue.
2) Wipe the exterior with a clean cloth; do not use cleaners. Use an alcohol swab if needed for the dirt removal.
B) Cartridge compartment cleaning:
1) Disconnect the Hemocue.
2) After opening the lid, rotate the cartridge compartment by pressing the 2 black knobs.
3) A sponge moistened with ethanol or any recommended disinfectants such as CaviCide should be used to wipe off the cartridge compartment. Avoid dripping any liquids inside.
C) Fan filter replacement: The fan filters located on the back should be regularly checked and replaced if clogged by dust particles. Extra filters are usually given with the analyzer or can be bought from the local distributor.
1) Remove the cover of the fan filter.
2) Remove the used fan filter and replace it with a new one if needed.
3) Cover the fan filter.
Choose two of the following tests/equipment below, tell us what test(s) can be performed and provide...
Problem 13. Refer to the triaxial shear test data (page below) to answer the following questions. a) What type of triaxial test was performed (CU, CD, or UU)? How can you tell? b) Samples of Soil 2 (see profile below) were obtained for the triaxial shear tests. Based on the results of the triaxial shear tests and the soil profile shown below, calculate the long-term shear strength of the soil at point A Soil 1: Soil 1 A 5 m...