Question

Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C between Oº and 200°C. The e

0 0
Add a comment Improve this question Transcribed image text
Answer #1

1. The uncertainity of measurement is range of values in which the true value lies from the measured value.
It is the sure range in which the true reading lies. All precise measurement instruments are callibrated and measurement uncertainity is given. It is given in various different ways like measure of extreme possible error, measure of minimum readable value etc.

For uncertainity of temperature measurement, the RTD reading is ±0.02°C and curve fit has error of 0.5°C.

Hence, max possible error = +0.02 + 0.5 = 0.52°C
Which means if thermocouple reads value of 100°C, the actual temperature will be 100±0.52 °C.

Thus, the uncertainty is ±0.52 °C. .........Answer 1


2. T = 0.11 + 19.2V
Given that V = 4mV = 0.004 V

T = 0.11 + 19.2 * 0.004  = 0.1868 °C

Thus, the temperature sensed by thermocouple will be 0.1868°C ..........Answer 2


Note: Please upvote if this helps you. In case of doubt, comment below. Good luck!

Add a comment
Know the answer?
Add Answer to:
Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C...
Your Answer:

Post as a guest

Your Name:

What's your source?

Earn Coins

Coins can be redeemed for fabulous gifts.

Not the answer you're looking for? Ask your own homework help question. Our experts will answer your question WITHIN MINUTES for Free.
Similar Homework Help Questions
  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C between 0 and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Experimental methodology Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to...

    Experimental methodology Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within £0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts....

  • class: experimental methots Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard...

    class: experimental methots Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15...

  • Problem 3 (30 pts.) A J-type thermocouple is calibrated against an RTD standard within 10.02°C between...

    Problem 3 (30 pts.) A J-type thermocouple is calibrated against an RTD standard within 10.02°C between 0 and 200°C. The emf is measured with a potentiometer having 0.005 mV resolution and an uncertainty of 0.045 mV over its operating range. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following equation: T(°C) = 0.11 + 19.16V - 0.06V2, where the units of voltage Vis mV. This second order curve fit has an error of...

  • experimental methodology - please solve ASAP Problem 3 (25 pts.) A J-type thermocouple is calibrated against...

    experimental methodology - please solve ASAP Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C....

ADVERTISEMENT
Free Homework Help App
Download From Google Play
Scan Your Homework
to Get Instant Free Answers
Need Online Homework Help?
Ask a Question
Get Answers For Free
Most questions answered within 3 hours.
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT