1. The uncertainity of measurement is range of values in which
the true value lies from the measured value.
It is the sure range in which the true reading lies. All precise
measurement instruments are callibrated and measurement
uncertainity is given. It is given in various different ways like
measure of extreme possible error, measure of minimum readable
value etc.
For uncertainity of temperature measurement, the RTD reading is
±0.02°C and curve fit has error of 0.5°C.
Hence, max possible error = +0.02 + 0.5 = 0.52°C
Which means if thermocouple reads value of 100°C, the actual
temperature will be 100±0.52 °C.
Thus, the uncertainty is ±0.52 °C. .........Answer 1
2. T = 0.11 + 19.2V
Given that V = 4mV = 0.004 V
T = 0.11 + 19.2 * 0.004 = 0.1868 °C
Thus, the temperature sensed by thermocouple will be 0.1868°C
..........Answer 2
Note: Please upvote if this helps you. In case of doubt, comment
below. Good luck!
Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C...
Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C between 0 and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...
Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...
Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...
Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...
Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...
Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...
Experimental methodology Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within £0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts....
class: experimental methots Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15...
Problem 3 (30 pts.) A J-type thermocouple is calibrated against an RTD standard within 10.02°C between 0 and 200°C. The emf is measured with a potentiometer having 0.005 mV resolution and an uncertainty of 0.045 mV over its operating range. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following equation: T(°C) = 0.11 + 19.16V - 0.06V2, where the units of voltage Vis mV. This second order curve fit has an error of...
experimental methodology - please solve ASAP Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C....