Question
class: experimental methots
Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 0.02°C between 0° and 200°C. The em
0 0
Add a comment Improve this question Transcribed image text
Answer #1

1. Uncertainity is range of values in which the true value lies. No measurement device is perfect. The accuracy for every instrument is limited. For example, while measuring length of a line, some may read it as 11.22mm while other 11.23 and third one 11.21. Now, we can surely say that the value is 11.2X . This X is nothing but the uncertainity. So, the measured value should be expressed as 11.22 ±0.01. Thus, rather measuring a fixed reading, we get a range of values in which the true value lies.

The RTD reading is ±0.02°C and curve fit has error of 0.5°C.

Hence, max possible error = +0.02 + 0.5 = 0.52

Thus, when we measure using this thermocouple as 50°C, the actual value will be 50±0.52 °C.

Thus, the uncertainty is ±0.52 °C. .........Answer 1


2. T = 0.11 + 19.2V
Given that V = 4mV = 0.004 V

T = 0.11 + 19.2 * 0.004  = 0.1868 °C

The thermocouple will sense a temperature of 0.1868°C ..........Answer 2


Note: Please upvote if this helps you. In case of doubt, comment below. Good luck!

Add a comment
Know the answer?
Add Answer to:
class: experimental methots Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard...
Your Answer:

Post as a guest

Your Name:

What's your source?

Earn Coins

Coins can be redeemed for fabulous gifts.

Not the answer you're looking for? Ask your own homework help question. Our experts will answer your question WITHIN MINUTES for Free.
Similar Homework Help Questions
  • Experimental methodology Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to...

    Experimental methodology Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within £0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts....

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C between 0 and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C between Oº and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • experimental methodology - please solve ASAP Problem 3 (25 pts.) A J-type thermocouple is calibrated against...

    experimental methodology - please solve ASAP Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C....

  • please help Pages 4 > of 5 Problem 3 (25 pts.) A J-type thermocouple is calibrated...

    please help Pages 4 > of 5 Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of...

ADVERTISEMENT
Free Homework Help App
Download From Google Play
Scan Your Homework
to Get Instant Free Answers
Need Online Homework Help?
Ask a Question
Get Answers For Free
Most questions answered within 3 hours.
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT