1. Uncertainity is range of values in which the true value lies.
No measurement device is perfect. The accuracy for every instrument
is limited. For example, while measuring length of a line, some may
read it as 11.22mm while other 11.23 and third one 11.21. Now, we
can surely say that the value is 11.2X . This X is nothing but the
uncertainity. So, the measured value should be expressed as 11.22
±0.01. Thus, rather measuring a fixed reading, we get a range of
values in which the true value lies.
The RTD reading is ±0.02°C and curve fit has error of 0.5°C.
Hence, max possible error = +0.02 + 0.5 = 0.52
Thus, when we measure using this thermocouple as 50°C, the actual
value will be 50±0.52 °C.
Thus, the uncertainty is ±0.52 °C. .........Answer 1
2. T = 0.11 + 19.2V
Given that V = 4mV = 0.004 V
T = 0.11 + 19.2 * 0.004 = 0.1868 °C
The thermocouple will sense a temperature of 0.1868°C
..........Answer 2
Note: Please upvote if this helps you. In case of doubt, comment
below. Good luck!
class: experimental methots Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard...
Experimental methodology Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within £0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts....
Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...
Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...
Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...
Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C between 0 and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...
Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...
Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C between Oº and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...
Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...
experimental methodology - please solve ASAP Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C....
please help Pages 4 > of 5 Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of...