Question

Problem 2 (25 pts.) A thermocouple is to be used with a 16 bit A/D converter that has an input range of ov to sv. The voltage
0 0
Add a comment Improve this question Transcribed image text
Answer #1

Given that A thermocouple is to be used with a 16 bit A/D converter that has an input range of OV to 5V. Given 16 bit A/D conThe lowest relative Quantization error is eRQ = Gain x eo = 101 % (-3.8146 x 10-5) ERQ = -3.8527mV (c) The indicated voltage

Add a comment
Know the answer?
Add Answer to:
Problem 2 (25 pts.) A thermocouple is to be used with a 16 bit A/D converter...
Your Answer:

Post as a guest

Your Name:

What's your source?

Earn Coins

Coins can be redeemed for fabulous gifts.

Not the answer you're looking for? Ask your own homework help question. Our experts will answer your question WITHIN MINUTES for Free.
Similar Homework Help Questions
  • 7 of 10 Problem 2 (25 pts) A thenmocouple is to be used with 16 bt A/D convener has full scale ra...

    7 of 10 Problem 2 (25 pts) A thenmocouple is to be used with 16 bt A/D convener has full scale rang of OV o SV. The volhage output of the themocouple range of S0'C to 70"C, the 3649 mV nespectively Parra) 5 pts Compute the quantiration enor and the temperature resolution for this confiration Part bi 10 pts An noninverting OP amp is to he used to amplify the voltage signal to reduce the relative varies lincarly with the...

  • Oil estion 2 (9%) The voltage output from a J-type thermocouple referenced 0°C is to be used to measure temperatures fr...

    Oil estion 2 (9%) The voltage output from a J-type thermocouple referenced 0°C is to be used to measure temperatures from 50 to 70°C. The output voltages vary linearly over this range from 2.585 to 3.649 mV 1. If the thermocouple voltage is input to a 12-bit A/D converter having a +5V range, estimate the percent quantization error in the digital value; 2. If the analog signal can be first passed through an amplifier circuit, compute the amplifier gain required...

  • Experimental methodology Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to...

    Experimental methodology Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within £0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts....

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C between 0 and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C between Oº and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

ADVERTISEMENT
Free Homework Help App
Download From Google Play
Scan Your Homework
to Get Instant Free Answers
Need Online Homework Help?
Ask a Question
Get Answers For Free
Most questions answered within 3 hours.
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT