A cardiac defibrillator is used to shock a heart that is beating erratically. A capacitor in this device is charged to 5.0 kV k V and stores 1100 J J of energy.
capacitance?
Given,
Potential difference ,V = 5000
Energy of the capacitor, E = 1/2*C*V² =1100 J
0.5*C*(5000)² = 1100 J
C = 2200/(5000)²
C = 0.000088 F = 88 microfarad is the correct answer
A cardiac defibrillator is used to shock a heart that is beating erratically. A capacitor in...
A cardiac defibrillator is used to shock a heart that is beating erratically. A capacitor in this device is charged to 5.0 kV and stores 1250 J of energy. What is its capacitance? Express your answer using two significant figures and include the appropriate units.
A cardiac defibrillator is used to shock a heart that is beating erratically. A capacitor in this device is charged to 5.0?kVand stores 1600J of energy. What is its capacitance? Express your answer using two significant figures and include the appropriate units. C =
A cardiac defibrilator is used to shock a heart that is beating erratically. A capacitor in this device is charged to 6000V and stores 182J of energy. What is its capacitance?
Cardiac defibrilator A cardiac defibrilator is used to shock a heart that is beating erratically. A capacitor in this device is charged to 5950 v and stores 195 J of energy. What is its capacitance? Submit Answer Tries 0/10
A defibrillator is used to restart a person's heart after it stops beating. Energy is delivered to the heart by discharging a capacitor through the body tissues near the heart. If the capacitance of the defibrillator is 8.000 µF and the energy delivered is to be 100.0 J, to what potential difference must the capacitor be charged? ___ kV
A defibrillator containing a 18.2 ?F18.2 μF capacitor is used to shock the heart of a patient by holding it to the patient's chest. Just prior to discharging, the capacitor has a voltage of 18.5 kV18.5 kV across its plates. How much energy is released into the patient? energy:
A defibrillator containing a 27.2 ?F capacitor is used to shock the heart of a patient by holding it to the patient's chest. Just prior to discharging, the capacitor has a voltage of 12.5 kV across its plates. How much energy is released into the patient?
A defibrillator containing a 14.2 μF capacitor is used to shock the heart of a patient by holding it to the patient's chest. Just prior to discharging, the capacitor has a voltage of 14.5 kV across its plates. How much energy is released into the patient?
A defibrillator passes a brief burst of current through the heart to restore normal beating. In one such defibrillator, a 59.5 µF capacitor is charged to 5.6 kV. Paddles are used to make an electrical connection to the patient's chest. A pulse of current lasting 1.0 ms partially discharges the capacitor through the patient. The electrical resistance of the patient (from paddle to paddle) is 226 Ω. (a) What is the initial energy stored in the capacitor? J (b) What...
A defibrillator passes a brief burst of current through the heart to restore normal beating. In one such defibrillator, a 40.6-μF capacitor is charged to 5.90 kV. Paddles are used to make an electric connection to the patient’s chest. A pulse of current lasting 1.00 ms partially discharges the capacitor through the patient. The electrical resistance of the patient (from paddle to paddle) is 240 Ω. What is the initial energy stored in the capacitor?