Consider a focused transducer with a radius of curvature of 10 cm and a diameter of 4 cm. This transducer operates at a frequency of 3.5 MHz, and transmits a pulse of duration 0.857 μs. What is the axial resolution, in mm, at the focal point of the transducer? If the axial resolution was to be improved by a factor-of-two for example, what physical or operating parameters could be changed? Explain.
solution::
please give me like.. its help me to write more questions..thank you..
Consider a focused transducer with a radius of curvature of 10 cm and a diameter of...
Consider a focused 1-cm-diameter transducer with 50% bandwidth in water. Its focal length is 2 cm. Along the transducer axis, there are two point scatters separated by 0.25 mm axially. Find the minimal central frequency of the transducer so that two scatters will appear in the ultrasound image?
10. Write a one-page summary of the attached paper? INTRODUCTION Many problems can develop in activated sludge operation that adversely affect effluent quality with origins in the engineering, hydraulic and microbiological components of the process. The real "heart" of the activated sludge system is the development and maintenance of a mixed microbial culture (activated sludge) that treats wastewater and which can be managed. One definition of a wastewater treatment plant operator is a "bug farmer", one who controls the aeration...