Patients with recurring ventricular arrhythmia, irregular beats of the heart's ventricles, are often fitted with implantable cardioverter defibrillators (ICD) that monitor the beating of the patient's heart. Much like an external defibrillator, an ICD delivers up to 40.0 J of energy to the heart when irregular heartbeat is detected. A standard ICD contains a 3.2 V lithium ion battery and has an equivalent capacitance of 140 µF.
a)
Charge
Q=CV =(140*10-6)(3.2) =4.48*10-4 C
b)
In a RC circuit charge as a function of time is given by
Q=Qmax(1-e-t/T)
0.932Qmax = Qmax (1-e-1.8/T)
Ln(1-0.932) = -1.8/T
T=0.66958 s
Since T=RC
R=0.66958/(140*10-6)
R=4782.72 ohms or 4.783 Kohms
c)
Time Constant
T=RC=500*140*10-6 = 0.07 s
For discharging
Q=Qmaxe-t/T
0.63Qmax = Qmaxe-t/0.07
Ln(0.63) =-t/0.07
t=0.03234 s
Patients with recurring ventricular arrhythmia, irregular beats of the heart's ventricles, are often fitted with implantable...
A heart dysfunction that can cause death is ventricular fibrillation. This is an uncoordinated quivering of the heart as opposed to regular beating. An electric shock to the chest can cause momentary paralysis of the heart muscle, after which the heart will sometimes start organized beating again. A defibrillator is a device that applies a strong electric shock to the chest over a time of a few milliseconds. The device contains a capacitor of several microfarads, charged to several thousand...
A defibrillator passes a brief burst of current through the heart to restore normal beating. In one such defibrillator, a 59.5 µF capacitor is charged to 5.6 kV. Paddles are used to make an electrical connection to the patient's chest. A pulse of current lasting 1.0 ms partially discharges the capacitor through the patient. The electrical resistance of the patient (from paddle to paddle) is 226 Ω. (a) What is the initial energy stored in the capacitor? J (b) What...
A defibrillator passes a brief burst of current through the heart to restore normal beating. In one such defibrillator, a 40.6-μF capacitor is charged to 5.90 kV. Paddles are used to make an electric connection to the patient’s chest. A pulse of current lasting 1.00 ms partially discharges the capacitor through the patient. The electrical resistance of the patient (from paddle to paddle) is 240 Ω. What is the initial energy stored in the capacitor?
The electrical signals to the heart are typically coordinated to create a wave of compression that travels across the heart and pumps the blood through the heart chambers. If these signals get misaligned, the bits of the heart muscle vibrate independently and as a result do not move blood. This situation is called fibrillation and can quickly lead to death. If the situation is caught in time, passing a jolt of current through the heart can realign the electrical signals...
Question 5 Part A Biological macromolecules such as proteins may be positively, negatively, or neutrally charged depending on what amino acids are present. Some parts of a macromolecule may be positively charged while others are negatively charged, and this can affect the conformation - the shape of the macromolecule. Changes in the conformation may affect the function of a macromolecule. Two charged macromolecules (labelled 1 and 2) have charges q1 =+3.20 * 10-19 C and q2 =+12.8 10-19 C. They...