Many studies have been conducted to determine the concentration of CO (carbon monoxide) near freeways with various conditions of traffic flow. The basic technique involves measuring CO concentrattions...

Many studies have been conducted to determine the concentration of CO (carbon monoxide) near freeways with various conditions of traffic flow. The basic technique involves measuring CO concentrattions with a spectrophotometer. These machines are quite delicate and have to be calibrated every day. If the machine reads close to 70 ppm on the span gas, it’s ready for use; if not, it has to be adjusted.
Each data set below represents some readings on span gas. Assume the Gaussian model, with errors following the normal curve. In each case, make a t-test to see whether the instrument is properly calibrated or not.
a) 71, 68, 79
b) 71, 68, 79, 84, 78, 85, 69
c) 71
d) 71, 84

0 0
Add a comment Improve this question Transcribed image text
Answer #1

To know, if the calibration is proper or not, we will run a 1 sample-t test

Null hypothesis: There is no significant difference between the mean reading of the machine and 70 ppm

Alternate hypothesis: There is a significant difference between the mean reading of the machine and 70 ppm

Readings
71
68
79
71
68
79
84
78
85
69
71
71
84
sample size (n) 13
Mean (µ) 75.2
Standard deviation (σ)(s) 6.5
df (n-1) 12

Assuming confidence level of 95%

with CI of 95% and df of 12, t value as per below table is 1.7832

Table of t-Distribution Areas df 0.25 0.20 0.15 0.10 0.05 0.025 0.02 0.01 0.005 0.0025 0.001 0.0005 1 1.000 1.376 1.963 3.078

We get below values

One-Sample T: Readings

Descriptive Statistics

N Mean StDev SE Mean 95% CI for μ
13 75.23 6.46 1.79 (71.33, 79.13)

μ: mean of Readings

Test

Null hypothesis H₀: μ = 70
Alternative hypothesis H₁: μ ≠ 70
T-Value P-Value
2.92 0.013

Since p-value is less than the level of significance, we reject the null hypothesis and conclude that there is a significant difference between the mean reading of the machine and 70 ppm

Hence the machine needs calibration

Add a comment
Know the answer?
Add Answer to:
Many studies have been conducted to determine the concentration of CO (carbon monoxide) near freeways with various conditions of traffic flow. The basic technique involves measuring CO concentrattions...
Your Answer:

Post as a guest

Your Name:

What's your source?

Earn Coins

Coins can be redeemed for fabulous gifts.

Not the answer you're looking for? Ask your own homework help question. Our experts will answer your question WITHIN MINUTES for Free.
Similar Homework Help Questions
ADVERTISEMENT
Free Homework Help App
Download From Google Play
Scan Your Homework
to Get Instant Free Answers
Need Online Homework Help?
Ask a Question
Get Answers For Free
Most questions answered within 3 hours.
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT