Question

Problem 3 (30 pts.) A J-type thermocouple is calibrated against an RTD standard within 20.02°C between 0º and 200°C. The emf

0 0
Add a comment Improve this question Transcribed image text
Answer #1

Given T(°C) = 0.11 + 19.16V - 0.06V2 a) The standard error of the fit for the second order polynomial is 0.45 Error contributemfo-25 + emf25-T1 = emfo-T1 This relationship allows the voltage reading from the nonstandard reference temperature to be co

Add a comment
Know the answer?
Add Answer to:
Problem 3 (30 pts.) A J-type thermocouple is calibrated against an RTD standard within 20.02°C between...
Your Answer:

Post as a guest

Your Name:

What's your source?

Earn Coins

Coins can be redeemed for fabulous gifts.

Not the answer you're looking for? Ask your own homework help question. Our experts will answer your question WITHIN MINUTES for Free.
Similar Homework Help Questions
  • Problem 3 (30 pts.) A J-type thermocouple is calibrated against an RTD standard within 10.02°C between...

    Problem 3 (30 pts.) A J-type thermocouple is calibrated against an RTD standard within 10.02°C between 0 and 200°C. The emf is measured with a potentiometer having 0.005 mV resolution and an uncertainty of 0.045 mV over its operating range. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following equation: T(°C) = 0.11 + 19.16V - 0.06V2, where the units of voltage Vis mV. This second order curve fit has an error of...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C between 0 and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 10.02°C between Oº and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C...

    Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within +0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts. Find the...

  • Experimental methodology Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to...

    Experimental methodology Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within £0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15 pts....

  • class: experimental methots Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard...

    class: experimental methots Problem 3 (25 pts.) A J-type thermocouple is calibrated against an RTD standard to within 0.02°C between 0° and 200°C. The emf is measured with an A/D converter having a resolution of 0.075 mV. The reference junction temperature is provided by an ice bath. The calibration procedure yields the following linear equation: T(°C) = 0.11 + 19.2V ,where the units of voltage Vis mV. This first order curve fit has an error of 0.5°C. Part a) 15...