Suppose that you have a time-of-flight mass spectrometer with a detector that can clearly resolve ion signs that are at least 14.0 nanoseconds apart in arrival time. Given a flight tube length of 1.40 m total field-free distance from source to detector, and a source with an acceleration voltage that extends over a 2.5 cm distance,
a) determine the source voltage required to ensure that two proteins with masses 2237.5 amu and 2239.8 amu are adequately resolved, i.e. have a flight time difference of 14 ns. Assume that all ions gain the full source energy possible, and that there is no temporal spread in the ion signal.
b) Assuming that the electric field is 2.400 kV m-1, determine the minimum length of field-free flight that would be required to resolve these two ions. (Note: Be sure to convert the electric field to a applied voltage!)
Suppose that you have a time-of-flight mass spectrometer with a detector that can clearly resolve ion...
An acceleration voltage of 30.0 kV is used to accelerate ions into the 1.54 m flight tube of a time‑of‑flight (TOF) mass spectrometer. The frequency at which spectra can be recorded is dependent on how long it takes the slowest ion to travel from the source to the detector. If you wish to scan up to ?/?m/z 6801 at what frequency (spectra/s) could spectra be recorded if a new spectrum begins each time the slowest ion reaches the detector? frequency:................spectra/s