Question

An acceleration voltage of 30.00 kV is used to accelerate ions into the 1.657 m flight tube of a ...

An acceleration voltage of 30.00 kV is used to accelerate ions into the 1.657 m flight tube of a time-of-flight mass spectrometer. If two singly, positively-charged ions, one with a m/z of 5068.664 and one with a m/z of 5077.857, are accelerated at the same time into the flight tube, what will be the difference in the arrival time of the two ions at the detector? Enter the difference in arrival time as a positive value.

Answer in ns.

0 0
Add a comment Improve this question Transcribed image text
Answer #1

as these are singly charged ions,

hence, charges of ions ( q1 =q2 ) = 1.6*10-19 C

m/z (mass to charge ratio) of 1st ion = 5068.664

hence, mass of 1st ion (m1 ) = 5068.664*mass of single charge = 5068.664 * 1.66*10-27 Kg =8.41398224*10-24 Kg

where, mass of single charge = 1.66*10-27 Kg

potential energy of 1st ion = q1 *V = 1.6*10-19 *30*103 J = 4.8*10-15 J

kinetic energy by this potential = 1/2*m1 *v12 = 4.8*10-15 J

hence,

0.5*8.41*10-24 *v12 = 4.8*10-15

v12 = 1.140957959*109

v1 = 3.37780692*104 m/s

time taken by 1st ion = length of tube / velocity of 1st ion = 1.657/ 3.37*104 s

t1 = 4.90554978*10-5 s -----(1)

similarly, for 2nd ion

m/z (mass to charge ratio) of 2nd ion = 5077.857

hence, mass of 2nd ion (m2 ) = 5077.857* mass of single charged = 5077.857*1.66*10-27 Kg =8.42924262*10-24 Kg

where, mass of single charged = 1.66*10-27 Kg

potential energy of 2nd ion = q2*V = 1.6*10-19 *30*103 J = 4.8*10-15 J

kinetic energy by this potential = 1/2*m2 *v22 = 4.8*10-15 J

hence,

0.5*8.43*10-24 *v22 = 4.8*10-15

v22 = 1.13889*109

v2 = 3.374747*104 m/s

time taken by 2nd ion = length of tube / velocity of 2nd ion = 1.657/ 3.374747*104 s

t2 = 4.909996 *10-5 s -----(2)

from 1 and 2

difference in arrival time = t2 -t1 =  4.909996 *10-5 s - 4.90554978*10-5 s = 4.4102*10-8 s = 44.102 ns

Add a comment
Know the answer?
Add Answer to:
An acceleration voltage of 30.00 kV is used to accelerate ions into the 1.657 m flight tube of a ...
Your Answer:

Post as a guest

Your Name:

What's your source?

Earn Coins

Coins can be redeemed for fabulous gifts.

Not the answer you're looking for? Ask your own homework help question. Our experts will answer your question WITHIN MINUTES for Free.
Similar Homework Help Questions
  • An acceleration voltage of 30.0 kV is used to accelerate ions into the 1.54 m flight tube of a time‑of‑flight (TOF) mass...

    An acceleration voltage of 30.0 kV is used to accelerate ions into the 1.54 m flight tube of a time‑of‑flight (TOF) mass spectrometer. The frequency at which spectra can be recorded is dependent on how long it takes the slowest ion to travel from the source to the detector. If you wish to scan up to ?/?m/z 6801 at what frequency (spectra/s) could spectra be recorded if a new spectrum begins each time the slowest ion reaches the detector? frequency:................spectra/s

  • Suppose that you have a time-of-flight mass spectrometer with a detector that can clearly resolve ion...

    Suppose that you have a time-of-flight mass spectrometer with a detector that can clearly resolve ion signs that are at least 14.0 nanoseconds apart in arrival time. Given a flight tube length of 1.40 m total field-free distance from source to detector, and a source with an acceleration voltage that extends over a 2.5 cm distance, a) determine the source voltage required to ensure that two proteins with masses 2237.5 amu and 2239.8 amu are adequately resolved, i.e. have a...

  • Resources hEUpt An electric field of 4000.0 V was used to accelerate a singly, positively-charged ion...

    Resources hEUpt An electric field of 4000.0 V was used to accelerate a singly, positively-charged ion with a m/z of 444.8 into the flight tube of a magnetic sector mass spectrometer. What is the velocity of the ion? m/s A homogenous magnetic field with a strength of 7.00 T is applied perpendicular to the length of the flight tube. What is the radius of curvature (in centimeters) for this ion in the magnetic field? cm. r=

  • In mass spectrometry, a sample’s molecules are ionized in a vacuum and then exposed to a...

    In mass spectrometry, a sample’s molecules are ionized in a vacuum and then exposed to a uniform electric field created by a high-voltage plate in the acceleration chamber. The electric field accelerates the ions until they arrive at the next section of the device, designated as the separation chamber. In this section, the drifting ions are sorted by their mass-to-charge ratio (m/q). The separation chamber in a time-of-flight mass spectrometer (TOF-MS) is linear and has no electric or magnetic fields....

  • Assuming the length of drift tube in ToF MS is 2.3 m and accelerating voltage is...

    Assuming the length of drift tube in ToF MS is 2.3 m and accelerating voltage is 30,000 V, one type of ions can reach the ion collector in 11.3 µs. What is the molar mass of the ions, assuming the ions are singly charged cations? Another instrumental analysis problem, please help!

  • 76. For a particular experiment, helium ions are to be given a kinetic energy of 3.0...

    76. For a particular experiment, helium ions are to be given a kinetic energy of 3.0 MeV. What should the voltage at the center of the ar celerator be, assuming that the ions start essentially at rest? A. -3.0 MV B. +3.0 MV C. +1.5 MV D. +1.0 MV s ti Materials analysis with ions. Rutherford backscattering spectrometry (RBS) is a technique used to determine the structure and composition of materials. A beam of ions (typically, helium ions) is accelerated...

  • Explain how a mass spectrometer is used to separate ions of different mass-to-charge (m/z) ratios by...

    Explain how a mass spectrometer is used to separate ions of different mass-to-charge (m/z) ratios by placing the statements in the order in which they occur. First step Gaseous ions are created in the ion source. The ions enter the vacuum chamber of the mass spectrometer A magnetic field is applied to the ions. The ions are deflected by the magnetic field. The velocity of the ions is increased as they pass through metal plates with a voltage applied to...

  • Constants Part A A mass spectrograph is used to measure the masses of ions, or to...

    Constants Part A A mass spectrograph is used to measure the masses of ions, or to separate ions of different masses. In one design for such an instrument, ions with mass m and charge q are accelerated through a potential difference V. They then enter a uniform magnetic field that is perpendicular to their velocity and are deflected in a semicircular path of radius R. A detector measures where the ions complete the semicircle and from this it is easy...

  • • The figure shows the essentials of a mass spectrometer, which can be used to measure...

    • The figure shows the essentials of a mass spectrometer, which can be used to measure the mass of an ion; an ion of mass m (to be measured) and charge q is produced in source S. The initially stationary ion is accelerated by the electric field due to a potential difference V. The ion leaves S and enters a separator chamber in which a uniform magnetic field is perpendicular to the path of the ion. A wide detector lines...

  • The picture above shows part of a mass spectrometer that can be used to measure molecular...

    The picture above shows part of a mass spectrometer that can be used to measure molecular charge-to-mass ratios. A charged molecule (orange circle) is ionized and accelerated through an electric potential difference into a region with a uniform magnetic field. Here the magnetic field points out of the screen. The field makes the positively charged molecules undergo circular motion as shown. By adjusting the voltage difference between the plates, one can change the radius of curvature of the charged particles...

ADVERTISEMENT
Free Homework Help App
Download From Google Play
Scan Your Homework
to Get Instant Free Answers
Need Online Homework Help?
Ask a Question
Get Answers For Free
Most questions answered within 3 hours.
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT