You perform a chemical reaction in a Styrofoam cup calorimeter (as described in the lab manual)...

You perform a chemical reaction in a Styrofoam cup calorimeter (as described in the lab manual) and measure an overall temperature change of 4.99 °C. The specific heat capacity of the solution is 4.18 J/(g °C), the heat capacity of the calorimeter is 15 J/°C, and the total solution mass is 48.3 g. How much heat energy (in kJ) was transferred to/from the cup and solution?

0 0
Add a comment Improve this question Transcribed image text
Answer #1

The answer is 1.08 kJ

Mass of solution, in = 48.39 Temperature change, at = 4.99€ Heat capacity of calorimeter, Ceal = 155/°C Specific heat capacit

Add a comment
Know the answer?
Add Answer to:
You perform a chemical reaction in a Styrofoam cup calorimeter (as described in the lab manual)...
Your Answer:

Post as a guest

Your Name:

What's your source?

Earn Coins

Coins can be redeemed for fabulous gifts.

Not the answer you're looking for? Ask your own homework help question. Our experts will answer your question WITHIN MINUTES for Free.
Similar Homework Help Questions
  • In the laboratory a "coffee cup" calorimeter, or constant pressure calorimeter, is frequently use...

    In the laboratory a "coffee cup" calorimeter, or constant pressure calorimeter, is frequently used to determine the specific heat of a solid, or to measure the energy of a solutiorn phase reaction. Thermometer Stirring rod A student heats 63.92 grams of iron to 98.03 °C and then drops it into a cup containing 75.92 grams of water at 24.47 °C. She measures the final temperature to be 30.66 °C The heat capacity of the calorimeter (sometimes referred to as the...

  • The In the laboratory a "coffee cup calorimeter, or constant pressure calorimeter, is frequently used to determi...

    The In the laboratory a "coffee cup calorimeter, or constant pressure calorimeter, is frequently used to determine the specific heat of a solid, or to measure the energy of a solution phase reaction. A student heats 61.44 grams of titanium to 97.72 °C and then drops it into a cup containing 83.69 grams of water at 21.75 °C. She measures the final temperature to be 27.96 °C. The heat capacity of the calorimeter (sometimes referred to as the calorimeter constant)...

  • chemical reaction was performed in a coffee cup calorimeter (the set up you experiment 2). The calorimeter containe...

    chemical reaction was performed in a coffee cup calorimeter (the set up you experiment 2). The calorimeter contained 50.0 g of water at an initial temperature After the reaction was complete the temperature of the water was 27.0°C. Coleute a. Calculate the energy change, in joules (J), of the water. The specific heats is 4.18 J/gºC. ater at an initial temperature of 22.0°C. ules ("), of the water. The specific heat (S.H.) of water Page 2 UTOROURUAN b. Identify the...

  • (5 pts.) A chemical reaction was performed in a coffee cup calorimeter (the set up you...

    (5 pts.) A chemical reaction was performed in a coffee cup calorimeter (the set up you used in experiment 2). The calorimeter contained 50.0 g of water at an initial temperature of 22.0°C. After the reaction was complete, the temperature of the water was 27.0 °C. a. Calculate the energy change, in joules (u), of the water. The specific heat (S.H.) of water is 4.18 J/gºC. b. Identify the reaction as endothermic or exothermic. Please explain why.

  • Thermometer When a solid dissolves in water, heat may be evolved or absorbed. The heat of aissolution (dissolving) can be determined using a coffee cup calorimeter. Cardboard or Styrofoam lid In the...

    Thermometer When a solid dissolves in water, heat may be evolved or absorbed. The heat of aissolution (dissolving) can be determined using a coffee cup calorimeter. Cardboard or Styrofoam lid In the laboratory a general chemistry student finds that when 11.13 g of CsBr(s) are dissolved in 114.50 g of water, the temperature of the solution drops from 25.61 to 22.35 °C The heat capacity of the calorimeter (sometimes refered to as the calorimeteγ constant) was determined in a separate...

  • When 0.0700 mol of a substance is added to 150.0 grams of an aqueous solution in...

    When 0.0700 mol of a substance is added to 150.0 grams of an aqueous solution in a Styrofoam cup calorimeter, a chemical reaction occurs that causes the temperature of the solution to increase from 25.3 to 38.3 °C. Assume that the specific heat of the solution is 4.18 J/(g°C) and that the heat capacity of the calorimeter is negligible. Use this data to calculate the enthalpy change (in kJ/mol substance) for the chemical reaction +116 kJ/mol +3.80 kJ/mol -571 kJ/mol...

  • A chemical reaction causes the temperature of the reservoir of a calorimeter containing 101.7 g of...

    A chemical reaction causes the temperature of the reservoir of a calorimeter containing 101.7 g of a dilute aqueous solution to rise from 22.6 °C to 27.7 °C. Calculate the heat q in joules transferred to the reservoir. Assume the specific heat capacity of the reservoir to be 4.18 J/g°C.

  • 4. A styrofoam-cup calorimeter containing 150 g of water is calibrated by generating 1770 J of...

    4. A styrofoam-cup calorimeter containing 150 g of water is calibrated by generating 1770 J of heat within it. The temperature increases by 1.67°C. What is the calorimeter constant Ccal (heat capacity of the calorimeter)? c(H20) = 4.184 Jg='K 5. A constant-pressure calorimeter, Where Ccal = 192 JK I. contains 100.0 mL of a 0.200 M aqueous solution of AgNO3. The temperature of the calorimeter and its contents is 22.30°C. Upon addition of 100.0 mL of 0.200 M solution of...

  • In one of the General Chemistry experiments, you used a coffee-cup calorimeter to measure the heat...

    In one of the General Chemistry experiments, you used a coffee-cup calorimeter to measure the heat of neutralization of selected acid-base reactions. The calorimeter consisted of two nested Styrofoam cups with a cardboard lid. A temperature probe was lowered in the solution through a hole in the lid. Is a coffee-cup calorimeter a constant-volume or a constant-pressure device? Is a coffee-cup calorimeter adiabatic? The coffee-cup calorimeter was calibrated by using the neutralization reaction between sodium hydroxide and hydrochloric acid. The...

  • In the laboratory a "coffee cup" calorimeter, or constant pressure calorimeter, is frequently used to determine...

    In the laboratory a "coffee cup" calorimeter, or constant pressure calorimeter, is frequently used to determine the specific heat of a solid, or to measure the energy of a solution phase reaction. Thermometer Stirring red A student heats 68.26 grams of titanium to 99.16 °C and then drops it into a cup containing 80.78 grams of water at 23.25 °C. She measures the final temperature to be 30.68 °C. The heat capacity of the calorimeter (sometimes referred to as the...

ADVERTISEMENT
Free Homework Help App
Download From Google Play
Scan Your Homework
to Get Instant Free Answers
Need Online Homework Help?
Ask a Question
Get Answers For Free
Most questions answered within 3 hours.
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT