Question

Sorry i was not given a charge in the problem, this is all the information i...

Sorry i was not given a charge in the problem, this is all the information i have!

A mass spectrometer is being used to separate common oxygen-16 from the much rarer oxygen-18, taken from a sample of old glacial ice. (The relative abundance of these oxygen isotopes is related to climatic temperature at the time the ice was deposited.) The ratio of the masses of these two ions is 16 to 18, the mass of oxygen-16 is 2.66 × 10-26 kg, and they are both singly charged and travel at 4.85 × 106 m/s in a 1.25 T magnetic field.

show answerNo Attempt   What is the separation between their paths in meters when they hit a target after traversing a semicircle?
Δd =

|

0 0
Add a comment Improve this question Transcribed image text
Answer #1

As we know that the separation in the paths is measured after traversing a semicircle, the separation is the difference in diameter (2r).

r=\frac{mv}{qB}

\Delta d=\frac{2(m_{18}-m_{16})v}{qB}

m_{18}-m_{16}=\frac{18-16}{18}\times 2.66\times 10^{-26}kg=3.33\times 10^{-27}kg

\Delta d=\frac{2(3.33\times 10^{-27}kg)(4.85\times 10^6)m/s}{(1.6\times 10^{-19}C)(1.25T)}

\Delta d=\frac{(3.23\times 10^{-20})}{(2\times 10^{-19})}

\Delta d=0.1615m

Add a comment
Know the answer?
Add Answer to:
Sorry i was not given a charge in the problem, this is all the information i...
Your Answer:

Post as a guest

Your Name:

What's your source?

Earn Coins

Coins can be redeemed for fabulous gifts.

Not the answer you're looking for? Ask your own homework help question. Our experts will answer your question WITHIN MINUTES for Free.
Similar Homework Help Questions
  • A mass spectrometer is being used to separate common oxygen-16 from the much rarer oxygen-18, taken...

    A mass spectrometer is being used to separate common oxygen-16 from the much rarer oxygen-18, taken from a sample of old glacial ice. (The relative abundance of these oxygen isotopes is related to climatic temperature at the time the ice was deposited.) The ratio of the masses of these two ions is 16 to 18, the mass of oxygen-16 is 2.66 × 10-26 kg, and they are both singly charged and travel at 5.05 × 106 m/s in a 1.15...

  • A mass spectrometer is being used to separate common oxygen-16 from the much rarer oxygen-18, taken...

    A mass spectrometer is being used to separate common oxygen-16 from the much rarer oxygen-18, taken from a sample of old glacial ice. The ratio of the masses of these two ions is 16 to 18, the mass of oxygen-16 is 2.66×10−26 kg, and oxygen-18 is 2.99 x10-26 kg they are negatively and singly charged and travel at 5.00×106 m/s in a 1.20-T magnetic field. What is the separation between their paths when they hit a target after traversing a...

  • please help me with question 5. Thank you Instructions: Showing Steps (or mathematics involved) is not...

    please help me with question 5. Thank you Instructions: Showing Steps (or mathematics involved) is not necessary until explicitly stated in the ques- tion. Just writing the final answer is fine. • However, if you decide to show steps, then you will get partial points for the steps even if the final answer is wrong. • Standard integrals can be written directly without showing the integration. • All final answers must be in SI units and the units must be...

  • Problem 12: Triply charged uranium-235 and uranium-238 ions are being separated in a mass spectrometer. (The...

    Problem 12: Triply charged uranium-235 and uranium-238 ions are being separated in a mass spectrometer. (The much rarer uranium-235 is used as reactor fuel.) The masses of the ions are 3.90 x 103 kg and 3.95 x 10 kg, respectively, and they travel at 3.1 x 10 m/s in a 0.22 T field. What is the separation between their paths in meters when they hit a target after traversing a semicircle? Numeric A numeric value is expected and not an...

  • (9%) Problem 5: Triply charged uranium-235 and uranium-238 ions are being separated in a mass spectrometer....

    (9%) Problem 5: Triply charged uranium-235 and uranium-238 ions are being separated in a mass spectrometer. (The much rarer uranium-235 is used as reactor fuel.) The masses of the ions are 3.90 x 10-25 kg and 3.95 x 10-25 kg, respectively, and they travel at 3.15 x 103 m/s in a 0.29 T field. * What is the separation between their paths in meters when they hit a target after traversing a semicircle? Ad = 0.2261 Grade Summary Deductions 30%...

ADVERTISEMENT
Free Homework Help App
Download From Google Play
Scan Your Homework
to Get Instant Free Answers
Need Online Homework Help?
Ask a Question
Get Answers For Free
Most questions answered within 3 hours.
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT