Question

In a magnetic sector mass spectrometer, would the measured mass increase or decrease if you: 1)....

In a magnetic sector mass spectrometer, would the measured mass increase or decrease if you:

1). Increase the accelerating potential

2). Decrease the magnetic field

Thank you for the help!

0 0
Add a comment Improve this question Transcribed image text
Answer #1

formula is given by

acceleration potential in the source region (V) and the mass to charge ratio (m/z) of the ion.

mass (m) and the velocity (v).

so from above formula it is clear  

m  ∝ B2   ( directly proportional )

and m  ∝ 1/v (inversely proportional)

so

1)

so if we increase the potential the mass will  decrease ( inversely proportional )

2)

and if we  Decrease the magnetic field the  mass will also  Decrease ( directly proportional )

Good Luck !!
let me know in a comment if there is any problem or doubts

Add a comment
Know the answer?
Add Answer to:
In a magnetic sector mass spectrometer, would the measured mass increase or decrease if you: 1)....
Your Answer:

Post as a guest

Your Name:

What's your source?

Earn Coins

Coins can be redeemed for fabulous gifts.

Not the answer you're looking for? Ask your own homework help question. Our experts will answer your question WITHIN MINUTES for Free.
Similar Homework Help Questions
  • How does changing the magnetic field strength in a magnetic sector mass spectrometer affect which ions...

    How does changing the magnetic field strength in a magnetic sector mass spectrometer affect which ions reach the detector? answers are: increase in magnetic field strength = ions with larger m/z reaching the detector, increase in magnetic field strength = ions with smaller m/z reaching the detector, or decrease in magnetic field strength = ions with larger m/z reaching the detector How does changing the accelerating voltage in a magnetic sector mass spectrometer affect which ions reach the detector? answers...

  • (Figure 1) shows a mass spectrometer, an analytical instrument used to identify the various molecules in...

    (Figure 1) shows a mass spectrometer, an analytical instrument used to identify the various molecules in a sample by measuring their charge-to-mass ratio q/m. The sample is ionized, the positive ions are accelerated (starting from rest) through a potential difference ΔV, and they then enter a region of uniform magnetic field. The field bends the ions into circular trajectories, but after just half a circle they either strike the wall or pass through a small opening to a detector. As...

  • The figure below shows a mass spectrometer, an analytical instrument used to identify the various molecules...

    The figure below shows a mass spectrometer, an analytical instrument used to identify the various molecules in a sample by measuring their charge-to-mass ratio e/m. The sample is ionized, the positive ions are accelerated (starting from rest) through a potential difference ΔV, and they then enter a region of uniform magnetic field. The field bends the ions into circular trajectories, but after just half a circle they either strike the wall or pass through a small opening to a detector....

  • In the Bainbridge mass spectrometer (Figure 1) , the magnetic-field magnitude in the velocity selector is...

    In the Bainbridge mass spectrometer (Figure 1) , the magnetic-field magnitude in the velocity selector is 0.510 T , and ions having a speed of 1.82×106m/s pass through undeflected. A) What is the electric-field magnitude in the velocity selector? b) If the separation of the plates is 5.20mm, what is the potential difference between the plates? Figure 1of 1 ㄧㄨㄨㄨ

  • In a mass spectrometer, a student needs to find the strength of the electric field that...

    In a mass spectrometer, a student needs to find the strength of the electric field that is needed in the velocity sector if a single charged particle with a mass of 7.971x10^-26 kg after leaving the velocity sector is to hit the ceiling 0.036 m above the opening where it enters the region with only magnetic field. There is no E field. The value of the magnetic field is 0.025 T in the entire mass spec. A) what is the...

  • AV An Iron-Nickel sample from a meteorite was placed into a velocity selector and mass spectrometer....

    AV An Iron-Nickel sample from a meteorite was placed into a velocity selector and mass spectrometer. Experimenters expect 3 primary isotopes: 56Fe, 58Ni, & boni. Positive ions are accelerated by an applied potential (AV) of 3300 Volts which is the same potential difference used to create the electric field in the velocity selector, where d is 1.1 cm. 1. On the figure provided, in the velocity selector region, (A) draw the direction of the electric field. (B) Draw the necessary...

  • An Iron-Nickel sample from a meteorite was placed into a velocity selector and mass spectrometer. Experimenters...

    An Iron-Nickel sample from a meteorite was placed into a velocity selector and mass spectrometer. Experimenters expect 3 primary isotopes: 56Fe, 58Ni, & 60Ni. Positive ions are accelerated by an applied potential (ΔV) of 3300 Volts which is the same potential difference used to create the electric field in the velocity selector, where d is 1.1 cm. Need 3 and 4. An Iron-Nickel sample from a meteorite was placed into a velocity selector and mass spectrometer. Experimenters expect 3 primary...

  • 10) Mass Spectrometer: (14 pts) A charged nucleus of mass m and charge q is accelerated...

    10) Mass Spectrometer: (14 pts) A charged nucleus of mass m and charge q is accelerated through a constant potential V and enters a region of constant magnetic field B. The nucleus then undergoes circular motion before hitting a detector. 10) Mass Spectrometer: (14 pts) A charged nucleus of mass m and charge q is accelerated through a constant potential V and enters a region of constant magnetic field B. The nucleus then undergoes circular motion before hitting a detector....

  • In a common type of mass spectrometer, a beam of ions is passed through a velocity...

    In a common type of mass spectrometer, a beam of ions is passed through a velocity sector with crossed electric and magnetic fields. What is the purpose of the velocity sector? to strip loose electrons from the ions to decrease the kinetic energy of the ions to prevent the ions from traveling in a circular path to block all ions except those with specific speeds

  • A beam of singly ionized uranium atoms 235 U and 238 U is injected into the mass spectrometer shown in the figure. The...

    A beam of singly ionized uranium atoms 235 U and 238 U is injected into the mass spectrometer shown in the figure. The ions all have the same velocities and charges. The 238 U ions follow the trajectory illustrated.1)Will the 235mU ions strike the collecting plate above, below, or at the same location as the 238mU ions?2)If the magnetic field strength in the spectrometer is increased, will the spacing between where the 235mU and 238mU ions strike increase, decrease, or stay...

ADVERTISEMENT
Free Homework Help App
Download From Google Play
Scan Your Homework
to Get Instant Free Answers
Need Online Homework Help?
Ask a Question
Get Answers For Free
Most questions answered within 3 hours.
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT