Question

Problems on Gel-Permeation Chromatography (GPC) 1. A gel-filteration column has a radius (r) of 0.80 cm and a length of 20.0

0 0
Add a comment Improve this question Transcribed image text
Answer #1

Answer 1):

a): Total volume of column= pi×r^2×l=3.14×(0.80^2)×(20.0) cm^3= 40.2 cm^3= 40.2 ml

b): Volume is occupied by the stationary phase plus the solvent inside the pores of the stationary phase= 18.2 ml + 40.2 ml= 58.4 ml

c): Volume (pores)= (40.2 ml× 60/100)=24.1 ml

Hence, volume range is 0-24.1 ml

{Please give a like}

Add a comment
Know the answer?
Add Answer to:
Problems on Gel-Permeation Chromatography (GPC) 1. A gel-filteration column has a radius (r) of 0.80 cm...
Your Answer:

Post as a guest

Your Name:

What's your source?

Earn Coins

Coins can be redeemed for fabulous gifts.

Not the answer you're looking for? Ask your own homework help question. Our experts will answer your question WITHIN MINUTES for Free.
Similar Homework Help Questions
  • 1.) Describe the goals of a Gel Filtration Chromotography Experiment??? 2.) Explain each key theoretical principle...

    1.) Describe the goals of a Gel Filtration Chromotography Experiment??? 2.) Explain each key theoretical principle of a Gel Filtration Chromotography, and how they help acheive the goal???. 4.) Explain the key equations used in the Gel Filtration Chromotography experiment and the terms involved in the equation???? HI im trying to prepare for a lab/report and i have some questions i could use help with please :) over all having trouble seeing how everything ties together etc :) thank you...

  • Chromatography Basics Multiple Choice: Answer the following multiple choice questions about Chromatography Basics  (SHOW WORKING): 1) In...

    Chromatography Basics Multiple Choice: Answer the following multiple choice questions about Chromatography Basics  (SHOW WORKING): 1) In a given chromatographic column, A) different solutes behave as if there are a different number of theoretical plates for each. B) the number of theoretical plates is the same for all solutes. C) efficiency is improved by increasing plate height. 2) What useful information can be found from a Van Deemter plot? A) The selectivity factor B) Optimum mobile phase flow rate C) Optimum...

  • 3. A capillary column for use in gas chromatography is 30.0 m long and has an...

    3. A capillary column for use in gas chromatography is 30.0 m long and has an internal diameter of 0.320 mm. The inside wall of the capillary is coated with the stationary phase. a. The phase ratio (B) is the ratio of the mobile phase volume to the stationary phase volume: B = Vm/Vs. For a particular separation, the desired phase ratio is B = 200. What should the thickness of the stationary phase be to obtain this phase ratio?...

  • 1-) Silica gel has a bulk density of about 0.3 g/cm3, so 15 g would occupy...

    1-) Silica gel has a bulk density of about 0.3 g/cm3, so 15 g would occupy a volume of 45-50 cm3 (45-50 mL). This quantity is called the: 2-) You are carrying out a chromatographic separation on a 15 mg sample using 6.5 g of silica gel in your column. If you aim for a column height of 10 cm of silica gel, what is the inside diameter of the necessary chromatography column in units of cm? 3-) Consider the...

  • 5. When all of the sample to be collected has left the column, will the drop...

    5. When all of the sample to be collected has left the column, will the drop of solution coming off the tip of the column be acidic, basic or neutral? 6. How many H+ ions would be displaced from the ion exchange column by 1 Cu2+ ion? 7. Binding sites on the ion exchange resin are saturated with which ion before the sample solution is added? 8. Ions are displaced from the resin binding sites by any ion that has...

ADVERTISEMENT
Free Homework Help App
Download From Google Play
Scan Your Homework
to Get Instant Free Answers
Need Online Homework Help?
Ask a Question
Get Answers For Free
Most questions answered within 3 hours.
ADVERTISEMENT
ADVERTISEMENT
ADVERTISEMENT